SOLVED A&d fx120i not weighing kernels or holding zero

Sokam101

Gunny Sergeant
Full Member
Minuteman
Mar 10, 2020
521
60
PROBLEM SOLVED. IT WAS USING THE ZERO TRACKING FEATURE AND IT WAS CORRECTING FOR THE .02 gn I KEPT ADDING

Is anyone any expert with this balance? I just got it in waiting for my auto trickler to come in.

It wont measure a few granules for me. It also loses zero after it does that. It goes up to .02 gn then it goes back to 0.00 after every drop

I followed all the instructions and I can't get it to stop doing this. It's level, it's away from other electronics, there is no air flow in the room, it's 68 degrees, Tried leaving it on for a day to warm up. I tried unplugging it.

Any other suggestions?
 
Last edited:
Is it possible that the slow trickle from empty pan is so slow that scale is thinking it is drift and automatically adjusting to maintain zero?

THAT WAS IT! THANK YOU SO MUCH!

It was set to 1. I turned it off and it started counting in .02 gn increments! What should it be set to when using the auto trickler?
 